test for iron- calibration curve and procedure



  • Hi,
    I have created a standard curve and procedure for the detection of iron with IO Rodeo. I had to do this for class, but I figured since I have all the data I might as well share it with the IO Rodeo community. I'm new to this world, so some help would be greatly appreciated. Below is a link to a Google Doc with the procedure and the xy data for the standard curve.

    https://docs.google.com/document/d/1JRNQIUFZDZik7C3rqL_tCDvc6tvb1Ny18nD0aygDDRo/edit?usp=sharing

    # 2018-05-17 19:19:59 PST
    # Colorimeter Data for iron (indicator 5x10-3 M 1,10-phenanthroline, add 5 mL to 100 mL of sample) 
    # linear
    # Hankard and Wolf, UST ’18
    # LED green 
    # -----------------------------
    # Concentration  | Absorbance
    0.00  0.00
    0.56  0.02
    1.40  0.04
    2.80  0.06
    4.20  0.10
    8.40  0.17
    16.8  0.33
    22.4  0.42
    