test for iron- calibration curve and procedure
-
Hi,
I have created a standard curve and procedure for the detection of iron with IO Rodeo. I had to do this for class, but I figured since I have all the data I might as well share it with the IO Rodeo community. I'm new to this world, so some help would be greatly appreciated. Below is a link to a Google Doc with the procedure and the xy data for the standard curve.https://docs.google.com/document/d/1JRNQIUFZDZik7C3rqL_tCDvc6tvb1Ny18nD0aygDDRo/edit?usp=sharing
# 2018-05-17 19:19:59 PST # Colorimeter Data for iron (indicator 5x10-3 M 1,10-phenanthroline, add 5 mL to 100 mL of sample) # linear # Hankard and Wolf, UST ’18 # LED green # ----------------------------- # Concentration | Absorbance 0.00 0.00 0.56 0.02 1.40 0.04 2.80 0.06 4.20 0.10 8.40 0.17 16.8 0.33 22.4 0.42